I might add that Fluke parts are usually available and relatively cheap. I heard that if you were to send in for repair, they calibrate it, but it is not included in the repair contract.

If you were to send in specifically for calibration, it is VERY expensive. I think $150 for the 87-V.
I've had Fluke throw in a FREE calibration on a warranty unit that they don't repair just replace and calibration wan't normally offer on the unit!

It wasn't a DMM though.
 

· Registered
Joined
·
250 Posts
Yep same here. the displayed values should not be relied on if the meter is not calibrated recently.
I can see getting a calibration after a repair and if you know readings are off but do you "actually" see any difference in readings with your annual calibration? I know your going to say that your job requires calibrated equipment, correct? Most don't fall into your catagory.

If you started a poll on how many electricians get their equipment calibrated annually, I might be wrong but I say very few do.
 