3yr apprentice
Joined
·
27 Posts
Discussion Starter · #1 ·
I'm going to be working on new dentist office and trying to confirm plug types I need for patient care are. It looks like it needs to be ISO ground. However some are saying it has to be hospital grade ISO ground. I don't see anything in code saying it HAS to be hospital grade. If I'm missing it. Let me know the code I'm not finding. Thanks
 

·
Electrical Simpleton
Joined
·
3,350 Posts
If it were NEC then hospital grade would not be required.

Pete
 

·
Donuts > Fried Eggs
Joined
·
17,042 Posts
Is anesthesia going to be used in the patient care area? If so, it looks to me like 517.61(C)(2) might apply and they would have to be hospital grade.

Otherwise I'd say it's not code, but could be spec.
 

·
Pool Shark
Joined
·
472 Posts
I know the NEC rules don't apply to you guys up there but I did a dental office and the line was drawn with local anesthesia and general anesthesia. If they were to practice general anesthesia I would've had to put in a backup power generator and lighting and fortify the medical operating room to medical grade specs.

However this wasn't the case for me here. Instead they were just doing local anesthesia. So I ran EMT with a redundant ground to the x-ray and sterilization equipment and supplied tamper resistant outlets for any outlet I thought a child could tamper with.

I also had to wire up a low voltage emergency cut off sensor to the nitrous oxide containers. Was fun but a was like pulling teeth:laughing: trying to get all the info I needed.
 

·
Registered
Joined
·
11,429 Posts
Is anesthesia going to be used in the patient care area? If so, it looks to me like 517.61(C)(2) might apply and they would have to be hospital grade.

Otherwise I'd say it's not code, but could be spec.
he is North of you, way north:whistling2::whistling2: oooh Canada
 

·
Registered
Joined
·
165 Posts
I'm not seeing anywhere that the code requires isolated ground for patient care areas. I'll be wiring a dentist office soon as we'll, and wouldn't mind a little clarity on this.
 

·
Electrical Contractor
Trying to retire or at least slow down a bit, but life not cooperating
Joined
·
5,062 Posts
I'm not seeing anywhere that the code requires isolated ground for patient care areas. I'll be wiring a dentist office soon as we'll, and wouldn't mind a little clarity on this.
Last dental office we did, the furniture came prewired to jbs in the kick space. We didn't have to do separate neutrals or full size grounds.

Get the engineer to clarify the area usage.
-
24-104 basically says you have a run a minimum #12 ground, and since they don't make #12 bare, you install green.
AND, you install a ground for EACH neutral you install.
Be careful when you get your iso ground #12 AC90, as they make one with the full size iso ground (that's the one you want) and the standard assembly which uses a#14 ground.
 

·
Registered
Joined
·
575 Posts
I'm not seeing anywhere that the code requires isolated ground for patient care areas. I'll be wiring a dentist office soon as we'll, and wouldn't mind a little clarity on this.

If you were in the states the NEC prohibits isolated ground receptacles in patient care areas.
 